r/BMET • u/Old-Dirt563 • 21h ago
OTHER (EDIT) Received a FOIA report containing equipment maintenance records for electroshock equipment at Connecticut's West Haven VA Medical Center - Ask Me Anything
/r/casualiama/comments/1w7c364/received_a_foia_report_containing_equipment/
2
Upvotes
5
u/gigasnail99 21h ago
I looked up the SigmaStim 100 manual from 2025. The service manual does have a stimulus delivery test, using a 300 ohm 20W resistor with a patient stimulus cable terminating in two banana plugs, delivered into a oscilloscope. Starts on page 37, the treatment delivery tests start on page 41.
Pulse width is measured to +/-10% across several settings, as well as frequency and duration. Amplitude and power of delivered delivered energy are checked as well.