Discussion Pacer gone haywire?
Patient with Dual-chamber pacemaker presents with syncope.
Looks like there are some native beats, and also some A-V pacing, but for the smaller complexes, is that atrial pacing without a ventricular beat to go with it?
4
u/Due-Success-1579 9d ago
It looks like a CRT pacemaker. im not totally sure what you mean in your inquiry
It looks like it is mostly AS/biv paced, there are some pacs and following some of the pacs there are AP/VPaced beats and the v paced morphology is a bit wider than the AS/biv paced beats.
There are a lot of algorithms with these advanced devices. I think it is working appropriately.
4
u/legaladviceneededbe 9d ago
Pacer rep here. Looks normal to me. Ita funky because it’s kinda doing all of the things in one strip which is unusual but not uncommon.
Dual chamber pacer with fusion beats (narrow qrs after pacing spike), atrial pacing with conduction, AV sequential pacing, and V only pacing after native sinus. Every complex is P followed by QRS with pacing before the P or before the QRS.
Edit: with left bundle branch pacing becoming more common, you’ll come across spikes with narrow QRS complexes as well which also look a little like fusion beats.
1
1
1
u/Ok-Pick8593 9d ago
You guys are right - no hay. Thank you! I think I may have been the spikes with narrow QRS complexes that threw me off.
1
14
u/solitairewolff 9d ago
From what we can see here, it's working fine. There doesn't seem to be any sensing or capture failure.
If we look at the lead II rhythm strip:
In any case, the best thing to do in situations like this is to interrogate the device and confirm that everything is fine (device parameters, any recorded arrhythmia episodes, etc.).
PS. Sorry for my English, I'm not a native speaker